06/23/09 - ICON supports the new Dräger C...
ICON extends 24x7 support to now include the Dräger Carina ventilator launched by Drä...
05/12/08 - ICON Becomes ANCC Accredited f...
ICON awarded certification as a Provider by ANCC
05/09/08 - ICON Attends NTI 2008
ICON supports Dräger and Gambro in NTI Exhibit Hall May 2008 May 6-8, 2008, Chicago...
04/01/08 - ICON launches live 24x7 suppor...
ICON now provides 24x7 live multi-professional support to Dräger customers in the UK ...
TopBar
 

Welcome Guest! Please log in or register.

|| Search || Members || Who's Online? ||
CRRT Forum > CRRT troubleshooting and alarms > Excess patient fluid loss/gain
 
Post New Topic Reply to this topic
  Subscribe to this topic Subscribe | Bookmark this topic Bookmark | Format topic for printing Print | Email topic to a friend Email << Previous Topic || Next Topic >>
Topic: Excess patient fluid loss/gain
Manny
New Member


Total Posts: 1
Joined: March 2008


Member Profile
Posted: March 18 2008 5:01 PM Email Manny Send a Private Message to Manny

Is the excess fluid loss/gain limit related to the patient fluid removal set flow rate?


Reply with a quote Edit this message Notify moderator  
Maureen
New Member


Total Posts: 1
Joined: March 2008


Member Profile
Posted: March 18 2008 6:13 PM Email Maureen Send a Private Message to Maureen

Hi Manny,

The 'excess patient fluid loss/gain limit' alarm is NOT related to the 'patient fluid removal (PFR) set flow rate. The Prisma Excess Patient fluid loss or gain is a user set limit that defines the maximum patient fluid variance (either a loss or gain) that is permissible within a 3 hour time window. The limit, which ranges from 130 to 400 mL is programmed during set up according to physician order or hospital protocol.  The limit cannot be adjusted during therapy.  This is an added layer of protection for patients. Once the limit is breached, the treatment must be terminated via 'end treatment' as opposed to changing the set. It is important to remember that the limit does not replace clinical judgment. Patient hemodynamic stability must be continuously a*sessed.

The PFR or 'patient fluid removal' is used for volume management and is the amount of fluid prescribed by the physician to be removed hourly by the Prisma pump.

Maureen



Reply with a quote Edit this message Notify moderator  
Paige
New Member


Total Posts: 2
Joined: March 2008
From: USA


Member Profile
Posted: March 26 2008 1:01 PM Email Paige Send a Private Message to Paige

Maureen,

Perhaps this is semantics, but I think that  the patient excess fluid limit is related to the PFR rate. The fluid balance applied toward the excess limit is the variance from the the run-time-adjusted actual patient fluid removal volume. If I set a PFR of 100 ml/hr (1.67 ml/min) and the pump, due to alarms and bag changes, only runs for 56 out of 60 minutes, then the time-adjusted PFR  would be 93 mL (56 minutes times 1.67 ml/min) . If the actual PFR is 73 ml then I would have a patient gain of 20 ml. The fluid balance is recalculated every minute over a running 3 hour window. Is this correct?


------------------