Topic: Drager XL and Nasal CPAP in NICU |
GJ RRT
Member
Total Posts: 73
Joined: September 2006
From: grand juncton, colorado
 |
Posted: April 10 2008 3:12 PM
Any issues with the XL and non-invasive nasal CPAP in NICU. Our NICU team is trying to get us to go with 100% XL's for NICU CPAP.
We currently use Cardinal's Airlife, and older infant flows. We trialed the Viasys SiPAP, but no one was convienced of it's benefit.
Thoughts, opinions, concerns,
GJ
------------------ Chris Hanson RN, RRT-NPS, CPFT, AE-C
ER Registered Nurse
Grand Junction, Colorado
|
Cowboy12
Member
Total Posts: 91
Joined: April 2002
From: Houston
 |
Posted: May 15 2008 10:08 PM 
Chris,
In my experience the biggest drawback to CPAP on the XL is loss of flow and volume monitoring, as the Neoflow sensor is removed.
Other than that, with appropriate prongs I have not had an issue with the XL in NIV.
Not much to it IMO – hope this is of some help